When measuring low resistances, test equipment operator technique becomes important, as test leads and connections can contribute a higher percentage of measuring error.
With digital meters, you have to subtract the reading made when the probe tips are touched together from the measured result.
Analog meters usually have a zero adjust which should be done on the very lowest scale.
Both meter types will have specification for accuracy. And, these accuracy limits must be added or subtracted to the measured result.
Imagine testing for .5 ohms with a meter that has an accuracy of plus or minus .5 ohms and a meter lead resistance of .4 ohms.
Cheers,